#179049 - What about the Termistor
Responding to: ???'s previous message
You may need a Platinum RTD

For any resistance device you can get rid of the non linearity by using a current source( or sink) rater then a resistor divider.
If you are running any kind of distance you need 4 wire kelvin sensing.

List of 24 messages in thread
TopicAuthorDate
Measure temperature @0.1deg resolution using thermistor            01/01/70 00:00      
   Sorry - doesn't work            01/01/70 00:00      
      Thermistors aren't such bad...            01/01/70 00:00      
         Remember the original requirement of 0.1°C            01/01/70 00:00      
   Resolution is not there...            01/01/70 00:00      
      What about the Termistor            01/01/70 00:00      
      RE: Resolution is not there...            01/01/70 00:00      
         one-wire sensors are still quite expensive than thermistors            01/01/70 00:00      
            Thermistor Vs one-wire sensor cost            01/01/70 00:00      
               Your processor has a voltage reference good enough?            01/01/70 00:00      
                  Voltage reference            01/01/70 00:00      
                     That voltage reference has large error...            01/01/70 00:00      
                  Vref is Not Needed            01/01/70 00:00      
                     Depends            01/01/70 00:00      
                        Ah but,            01/01/70 00:00      
                           Still not trivial            01/01/70 00:00      
         The normal fluke doesn't give 0.1C correct readings            01/01/70 00:00      
         I think your customer ...            01/01/70 00:00      
            quite common problem with resolution/precision            01/01/70 00:00      
         linearing only solves one issue            01/01/70 00:00      
            Precision            01/01/70 00:00      
               hm            01/01/70 00:00      
               Not an opamp            01/01/70 00:00      
               Comparison with Fluke a big problem            01/01/70 00:00      

Back to Subject List