#181456 - That's great! (but some other concerns...)
Responding to: ???'s previous message
Kai,
That really cheered me up!

The EMG signals are not continuous signals but rather intermittent pulse trains( When the patient activates his muscle, a single Gaussian pulse train of the said frequency is generated.) So I don't know whether that would be a problem for the RMS calculation for the AD637.
But I'm looking into the datasheet for other possible hurdles that you mentioned.

Finally, since I haven't worked with these before, my ultimate doubt is whether hardware RMS be a better option than a software solution? (It would take some burden off the code I'm writing to process the EMG signals, but I'm wondering whether the accuracy would be better.)

Thanks a lot.

List of 28 messages in thread
TopicAuthorDate
Adding a DC biac to a bipolar signal?            01/01/70 00:00      
   summing            01/01/70 00:00      
      Resitor bias            01/01/70 00:00      
         Yes, mentioned but maybe missed            01/01/70 00:00      
   What about the character of the signal?            01/01/70 00:00      
      Information content..            01/01/70 00:00      
         Hardly....            01/01/70 00:00      
            So the only option would be software RMS?            01/01/70 00:00      
               100Hz or 100 beats per minute?            01/01/70 00:00      
                  100Hz.            01/01/70 00:00      
                     50-150Hz is in the range of AD637!            01/01/70 00:00      
                        That's great! (but some other concerns...)            01/01/70 00:00      
                           Can you show us a typical signal?            01/01/70 00:00      
                              A Typical EMG Signal            01/01/70 00:00      
                                 Probably tough for rms chi            01/01/70 00:00      
                                    I'm working with a dsPIC            01/01/70 00:00      
                                       Shouldn't be too expensive            01/01/70 00:00      
                                 Give the AD637 a try...            01/01/70 00:00      
                                    Look at LTC1966            01/01/70 00:00      
                                       Interesting...            01/01/70 00:00      
                                          Just received a fresh new batch of AD637s            01/01/70 00:00      
               Negative supply voltage available?            01/01/70 00:00      
   Does your system have a driven electrode?            01/01/70 00:00      
   better be careful            01/01/70 00:00      
      Regrowth            01/01/70 00:00      
         are you suggesting that humans are expendable?            01/01/70 00:00      
            All precautions taken.            01/01/70 00:00      
            Obviously            01/01/70 00:00      

Back to Subject List